I'm going to make this a FDA alert post for all the FDA alerts. I will post all the alerts here and if you come across any that I dont post them post them here also. Then we will have them all together. I've asked the moderator to make it a sticky so it will stay at the top. Some are serious alerts and NEED to be available for all to see.
Blessings,
Gina Marie

GinaMarie
01-05-2009, 12:53 PM
http://www.fda.gov/oc/po/firmrecalls/stryker12_08.html

Recall -- Firm Press Release

FDA posts press releases and other notices of recalls and market withdrawals from the firms involved as a service to consumers, the media, and other interested parties. FDA does not endorse either the product or the company.

Stryker Craniomaxillofacial Issues Nationwide Recall of its Custom Cranial Implant Kits; Catalogue Numbers 54-00101, 54-00102, 54-00103 and 54-00104, all lots shipped sterile

Contact:
Patrick Anderson, Stryker Corporation
(269) 385-7253
Patrick.Anderson@Stryker.com

Aaron Kwittken, Kwittken & Company
(646) 747-7144
AKwittken@Kwitco.com

FOR IMMEDIATE RELEASE -- Portage, Mich., -- Dec. 23, 2008 --- Stryker's Craniomaxillofacial (CMF) business unit announced today that the U.S. Food and Drug Administration (FDA) concluded on Dec. 18, 2008 that its recall of the following Stryker CMF's Custom Cranial Implant Kits is a Class I recall, which means the product could pose an imminent hazard to health: Catalogue Numbers 54-00101, 54-00102, 54-00103 and 54-00104, all lots shipped sterile.

The notification relates to a nationwide voluntary recall initiated by Stryker CMF on Oct. 24, 2008 of 322 Custom Cranial Implant Kits after determining that the sterilization validation of the product was not performed according to appropriate standards. Because the company could not assure the sterility of the product, implantation could result in serious health problems including infection (e.g., meningitis), intracranial abscess, wound infection, sepsis, re-operation with revision of cranioplasty, long-term neurological deterioration, additional hospitalization, or long-term intravenous antibiotic treatment. Death can occur if an infection is not diagnosed quickly.

Physicians and hospitals who have product that corresponds to the catalogue numbers listed above should stop implanting the product immediately and return it to Stryker CMF. Stryker recommends that patients who have received these implants be monitored by their physicians for signs of infection for at least six months post-implantation.

Stryker CMF previously notified implanting surgeons, hospital risk managers and its sales representatives of the potential health risks associated with the kits and requested that they return any remaining products to Stryker CMF. Stryker CMF will again notify implanting surgeons, risk managers and its sales representatives of this recall and its potential health risk and request that physicians or hospitals return to Stryker CMF any of this product that has not been implanted.

Physicians and patients with questions related to this issue should contact Stryker CMF at 800-962-6558, Monday – Friday, 8 a.m. to 7 p.m. ET. Patients with questions are encouraged to speak with their surgeon. Any adverse events relating to this product should be reported to Stryker.

Any adverse reactions experienced with the use of this product, and/or quality problems should also be reported to the FDA's MedWatch Program by phone at 1-800-FDA-1088, by Fax at 1-800-FDA-0178, by mail at MedWatch, HF-2, FDA, 5600 Fishers Lane, Rockville, MD 20852-9787, or on the MedWatch website at www.fda.gov/medwatch.

Stryker Corporation is one of the world's leading medical technology companies with the most broadly based range of products in orthopaedics and a significant presence in other medical specialties. For more information about Stryker, please visit www.stryker.com.

#

RSS Feed for FDA Recalls Information [what's this?]

GinaMarie
01-05-2009, 12:54 PM
http://www.fda.gov/bbs/topics/NEWS/2009/NEW01937.html

FDA News
FOR IMMEDIATE RELEASE

January 2, 2009
Media Inquiries:
Siobhan DeLancey, 301-796-4668
Consumer Inquiries:
888-INFO-FDA



FDA Announces Class I Recall of Ophthalmic Surgical Device

The U.S. Food and Drug Administration announced a Class I recall of lot no. UD30654 of Healon D, an ophthalmic viscosurgical device (OVD) manufactured by Advanced Medical Optics Inc. (AMO) of Santa Ana, Calif.

OVDs are viscoelastic materials used to maintain space in the eye during surgery. Typically, OVDs are pre-packaged in a syringe and are applied using a small tube.

On Oct. 30, 2008, AMO voluntarily recalled all 4,439 units of Healon D lot no. UD30654 of Healon D, due to complaints of inflammation after eye surgery, including Toxic Anterior Segment Syndrome (TASS). At that time, AMO informed customers of the number and nature of adverse event reports associated with OVD from that lot, and included a fax reply form for quick communication.

However, as of December 3, 2008, AMO had retrieved only 964 units of the 1,450 that had been distributed in the United States.

AMO received 66 adverse event reports associated with the recalled products. Tests of this lot revealed elevated levels of endotoxin, which has been associated with post-operative intraocular inflammation and TASS.

TASS is a post-operative, acute inflammation of the anterior segment of the eye (the front third of the eye including the cornea, iris and lens). TASS has been linked to solutions and devices used during eye surgery, such as OVDs, intraocular lenses and irrigation solutions.

The FDA urges anyone in possession of any units from the recalled lot, whether units were purchased from the company or provided as a sample by a sales representative, to remove them from inventory, and contact AMO at 1-877-AMO-4Life to make arrangements for return. The lot number of the device is displayed on the side panel of each unit.

Health care professionals and consumers may report serious adverse events (side effects) or product quality problems with the use of this product to AMO by calling 1-877-AMO-4LIFE and to the FDA’s MedWatch Adverse Event Reporting program either online, by regular mail, fax or phone.

Online: www.fda.gov/MedWatch/report.htm
Regular Mail: use postage-paid FDA form 3500 available at www.fda.gov/MedWatch/getforms.htm and mail to MedWatch, 5600 Fishers Lane, Rockville, MD, 20852-9787
Fax: (800) FDA-0178
Phone: (800) FDA-1088

GinaMarie
01-05-2009, 01:29 PM
http://www.fda.gov/oc/po/firmrecalls/ethex12_08.html

Recall -- Firm Press Release

FDA posts press releases and other notices of recalls and market withdrawals from the firms involved as a service to consumers, the media, and other interested parties. FDA does not endorse either the product or the company.

ETHEX Corporation Initiated Nationwide Voluntary Recall of a Single Lot of Hydromorphone HCl 2 mg Tablets Due to Potential for Oversized Tablet

Contact:
Ann McBride
1-800-748-1472

FOR IMMEDIATE RELEASE -- St. Louis, MO – December 23, 2008 – ETHEX Corporation announced today that it has voluntarily recalled to the consumer level, a single production lot of Hydromorphone HCl 2 mg tablets (Lot #90219, Exp: 03/2010; NDC #58177-0620-04), as a precaution, due to the possibility it may contain oversized tablets. Hydromorphone is a drug used for pain management and is packaged under the ETHEX label in 100-count bottles.

If someone were to take a higher than expected dose of Hydromorphone, the risk of adverse effects known to be associated with the drug may be increased, including respiratory depression (difficulty or lack of breathing), low blood pressure, and sedation.

There are other companies in the United States producing and marketing versions of Hydromorphone HCl tablets and consumers and their caregivers are encouraged to check their prescriptions to determine the source of their tablets. Hydromorphone HCl 2 mg tablets marketed by ETHEX are a blue, round tablet with a script "E" on one side and a "2" on the other side.

ETHEX Corporation has initiated recall notifications to *********rs and retailers nationwide who have received any inventory of the recalled lot of this product with instructions for returning the recalled product and, if they have not already done so, they are urged to contact ETHEX as provided below regarding procedures for returning the recalled product. If consumers have any questions about the recall, they should call the telephone number below or their physician, pharmacist, or other health care provider.

Any customer inquiries related to this action should be addressed to ETHEX Customer Service at 1-800-748-1472 or fax to ETHEX Customer Service at 314-646-3751, or e-mail to customer-service@ethex.com. Representatives are available Monday through Friday, 8 am to 5 pm CST. Consumers who experience any adverse reactions to this drug should contact their physician and/or healthcare provider immediately. Any adverse reactions experienced with the use of this product, and/or quality problems may also be reported to the FDA's MedWatch Program by phone at 1-800-FDA-1088, by fax at 1-800-FDA-0178, by mail at MedWatch, FDA, 5600 Fishers Lane, Rockville, MD 20852-9787, or on the MedWatch website at www.fda.gov/medwatch.

The Hydromorphone HCl recall announcement is posted on www.kvpharmaceutical.com and www.fda.gov/opacom/7alerts.html. It includes step-by-step details on how to return affected product to KV Pharmaceutical. The Company web site also includes a list of the drugs affected by the suspension.

The parent company of ETHEX Corporation, KV Pharmaceutical has advised the U.S. Food and Drug Administration that, effective midnight Dec. 19, 2008, the company voluntarily suspended shipments of all FDA-approved drug products in tablet form. This action is being taken as a precautionary measure, to allow KV to expeditiously address manufacturing issues that have come to management’s attention, to review and enhance comprehensively the company’s quality systems, and to implement efficiency improvements in its production facilities. KV is keeping the FDA informed about the Company’s plans.

This recall and suspension are being conducted with the knowledge of the FDA. At this time, the company is unable to determine when distribution of tablet form products will resume, or estimate what the financial impact of the recall and suspension will be.

GinaMarie
01-05-2009, 01:57 PM
http://www.fda.gov/oc/po/firmrecalls/national12_08.html

Recall -- Firm Press Release

FDA posts press releases and other notices of recalls and market withdrawals from the firms involved as a service to consumers, the media, and other interested parties. FDA does not endorse either the product or the company.

National Brands Inc. Issues a Nationwide Recall of Topaz brand Wafer Rolls Because of Possible Health Risks

Contact:
National Brands Inc.
1-866-238-5201

FOR IMMEDIATE RELEASE -- December 19, 2008 – National Brands Inc., Spring Valley NY is initiating a nationwide recall of all their 4.76oz (135gm) cans and 12.3oz (350gm) cans of Topaz brand Wafer Rolls because it may be contaminated with Melamine.

Product was distributed nationwide through retail stores.

The product comes in 4.76oz (135gm) metal cans and 12.3oz (350gm) metal cans in the following four flavors:

*Topaz Wafer Rolls with Chocolate Flavored Cream Filling Lot numbers L8085A, L8219A, L8245A

*Topaz Wafer Rolls with Hazelnut Chocolate Flavored Cream Filling Lot numbers L8085D, L8219D, L8245D

*Topaz Wafer Rolls with Vanilla Flavored Cream Filling Lot numbers L8085B, L8219B, L8245B

*Topaz Wafer Rolls with Mocha Cappuccino Flavored Cream Filling Lot numbers L8085C, L8219C, L8245C

Consumers who have the product which is being recalled should stop using it immediately. No illnesses associated with this product have been reported to date.

The recall was initiated after FDA testing discovered that product was found to contain Melamine.

Consumers who have purchased Topaz brand Wafer Rolls are urged to return it to the place of purchase for a full refund.

Consumers with questions may contact the company at 1-866-238-5201, Monday to Friday 9:00 to 5:00, Eastern Standard Time.

GinaMarie
01-07-2009, 08:32 PM
Recall -- Firm Press Release

FDA posts press releases and other notices of recalls and market withdrawals from the firms involved as a service to consumers, the media, and other interested parties. FDA does not endorse either the product or the company.

Celeste Industries Corporation Recalls All Lots of simplySmart Remove Make Up Remover

Contact:
Celeste Industries Corporation
410-822-6200, ext 1349

FOR IMMEDIATE RELEASE -- Easton, Maryland – January 5, 2009 – Celeste Industries Corporation has voluntarily recalled all lots of simplySmart™ "Remove" Make Up Remover in the US and Canada. This product is packaged as a single towelette with hiexpress.com shown on the packet. This recall was a result of certain limited lots testing positive for Pseudomonas aeruginosa bacteria.

The organism Pseudomonas aeruginosa may cause serious eye infections, respiratory infections, dermatitis, soft tissue infections, bacteremia, and a variety of systemic infections, particularly in patients who are immunosuppressed. Because the Make-Up Remover may be used in the area of the eye, there is a possibility that inadvertent introduction of the tainted product directly in the eye could result in serious eye infections.

No illnesses have been reported to date in connection with this problem. The issue was discovered after a consumer complained of an odor coming from the towelette packets.

This product is not sold in the retail market but is provided as a guest amenity at Holiday Inn Express® hotels in the United States in Canada.

Celeste Industries ceased production of the product on October 5, 2008 and subsequently notified the distributor and the Food and Drug Administration

Any hotel guests who may have taken the product are advised to not use the product and to destroy it immediately. Consumers who have used the product and have any concerns should contact a healthcare professional.

Celeste Industries has identified the source of the contamination and will take all necessary steps to prevent future contamination when production of the product is resumed.

Consumers with questions regarding the recall should contact Celeste Industries Corporation at 410-822-6200, ext 1349 between 8:00 a.m. and 5:00 p.m. EST and ask for Quality Assurance or e-mail at recall@celestecorp.com. This recall is being made in cooperation with the Food and Drug Administration.